Solved

Director of Testing Laboratory

Posted on 2016-09-22
4
47 Views
Last Modified: 2016-10-14
We recently purchased an ICP instrument from thermo-fisher.  My question has to do with the Argon gas supply.  We purchased one of the very large Argon dewers, and it it lasted for only two weeks.  I was told this has to do with the venting of the dewer, and that this short life was expected.  Is it better to buy the smaller cylinders that do not vent?
I was also told that they recommend keeping the instrument in stand-by mode, and thus constantly Argon purged for the sake of the internal optics of the instrument.  Any experience out there?  Need to get this resolved quickly.
0
Comment
Question by:Daniel Knauer
[X]
Welcome to Experts Exchange

Add your voice to the tech community where 5M+ people just like you are talking about what matters.

  • Help others & share knowledge
  • Earn cash & points
  • Learn & ask questions
  • 2
4 Comments
 
LVL 27

Assisted Solution

by:d-glitch
d-glitch earned 150 total points (awarded by participants)
ID: 41811413
Do you have a model number for the instrument?

Does it require liquid argon or dry gas?  
I think all dewars need to vent.  Are compressed gas bottles an option?
How long would a standard gas bottle last?  Have you done those calculations?
Have you talked to the manufacturer about your requirements, including an operating scenario, and availability?

What are the consequences of not purging?  Is the instrument destroyed or does it require some warm-time or recalibration when you bring it back on line?
0
 
LVL 81

Accepted Solution

by:
byundt earned 350 total points (awarded by participants)
ID: 41815121
I know absolutely nothing about inductively coupled plasma instruments, but found some specs for an instrument on line and did some calcs its Argon consumption. If my naive understanding is correct, you would be far better off with a gas cyclinder.

A Thermo Fisher model iCAP RQ ICP-MS (bench top unit) uses a maximum of 10 mL/min of Argon for cell gas supply. If you keep the system in standby mode, that is a consumption of 0.6 L/hr. A size T cylinder holds 9290 L of Argon at STP, while a size K cylinder holds 6880 L. Those correspond to 92 and 68 weeks at 10 mL/min, respectively.
0
 
LVL 27

Expert Comment

by:d-glitch
ID: 41843334
Reasonable research and suggestions for this abandoned question.
0

Featured Post

[Webinar] How Hackers Steal Your Credentials

Do You Know How Hackers Steal Your Credentials? Join us and Skyport Systems to learn how hackers steal your credentials and why Active Directory must be secure to stop them. Thursday, July 13, 2017 10:00 A.M. PDT

Question has a verified solution.

If you are experiencing a similar issue, please ask a related question

Digital marketing agencies have encountered both the opportunities and difficulties that emerge from working with a wide-ranging organizations.
This paper addresses the security of Sennheiser DECT Contact Center and Office (CC&O) headsets. It describes the DECT security chain comprised of “Pairing”, “Per Call Authentication” and “Encryption”, which are all part of the standard DECT protocol.
The Bounty Board allows you to request an article or video on any technical topic, or fulfill a bounty request to earn points. Watch this video to learn how to use the Bounty Board to get the content you want, earn points, and browse submitted bount…
Articles on a wide range of technology and professional topics are available on Experts Exchange. These resources are written by members, for members, and can be written about any topic you feel passionate about. Learn how to best write an article t…

705 members asked questions and received personalized solutions in the past 7 days.

Join the community of 500,000 technology professionals and ask your questions.

Join & Ask a Question